Atrium IIoT Gateway › AssetPulse Lite
Atrium App
AssetPulse Lite User Guide
AssetPulse Lite collects a raw vibration waveform from a sensor when a reading crosses a level you set, and stores it on the gateway for you to download or pull into your own analysis software. Everything it does can also be set up and read over HTTP.
Audience
Reliability engineers and integrators
Requires
Atrium 2.9.3 or newer
License
Not required
01Overview
AssetPulse Lite decides when a vibration sensor’s raw waveform is worth collecting, asks the sensor for it, and records why.
NCD vibration sensors sleep between reports. Each report carries summary readings (velocity, RMS acceleration, peak acceleration and temperature), and for about two seconds afterwards the sensor listens. AssetPulse Lite checks every report from a sensor it watches. If a threshold says so, it requests the raw time-domain waveform inside that listening window, so the waveform is of the machine in the state that produced the reading.
The app does no analysis of its own. It does not calculate an FFT, grade severity or write a report. It collects waveforms and hands them over. For analysis on the gateway, see AssetPulse.
A waveform per run
Set a Waveform Collection level between a machine’s idle and running readings, and a waveform is collected while it runs, as often as you choose.
A Critical level
A second, higher level collects a waveform straight away, whatever was collected earlier, and can email the people who need to know.
Two-probe sensors
A two-channel sensor is one entry with thresholds per probe, and one request can collect both probes.
Built for your own software
Every setting and every waveform is available over an HTTP REST API with an API token, as well as CSV download and MQTT.
assetpulse lite atrium app guide 01 sensors
AssetPulse Lite › Sensors
02Requirements
What has to be in place before you add your first sensor.
ItemRequirement
Atrium platform2.9.3 or newer.
GatewayAny Atrium gateway model, including the EG5100, and the container deployment.
SensorsOne of the vibration sensors below, whitelisted on the gateway and reporting.
One gateway per sensorA sensor should be managed by one gateway only. Two gateways managing the same sensor send it conflicting settings and cause waveform requests to fail.
Email (optional)A mail server set under the gateway’s Settings › SMTP, if you want Critical alert emails.
LicenseThe app is free and needs no license.
Supported sensors
03Installing
Install the app from the gateway’s Apps page like any other Atrium app.
  1. 1
    Open Apps on the gateway dashboard
    Click Browse App Center and install AssetPulse Lite, or click Install App and upload the package file.
  2. 2
    Wait for the services to restart
    Installing restarts the gateway’s services, which takes a few seconds. The app then appears on the Apps page as AssetPulse Lite.
  3. 3
    Open the app
    Click Open on the card. The app opens on the Sensors screen, which is empty until you add a sensor.
assetpulse lite atrium app guide 02 apps page
Apps
04First-Time Setup
The shortest path to a collected waveform: add a sensor, set one level, and watch the first capture arrive.
1. Add a sensor
  1. 1
    Click Add sensor
    On the Sensors screen, click Add sensor.
  2. 2
    Choose the sensor
    Under Vibration sensor, pick the sensor. Only whitelisted vibration sensors that are not already added are listed. A two-probe sensor is listed once and marked 2 probes.
  3. 3
    Describe it (optional)
    Enter a Name, what it is Installed on and its Location. These start from what is already entered on the gateway’s own sensor page, and all three are optional.
  4. 4
    Click Add sensor
    Leave Minimum time between collections at 0 for now. The sensor’s page opens. Nothing is collected until a threshold is switched on.
2. Set the collection level
  1. 1
    Read the Velocity chart
    On the sensor’s page, find the Velocity panel. The orange line is the highest reading in each interval and the grey line the lowest. On a machine that starts and stops, the low readings sit at the idle value and the high readings at the running value.
  2. 2
    Enter a level between idle and running
    Under Waveform Collection, enter a level in Collect at that is above the idle value and below the running value, and switch the threshold on. A dashed line appears on the chart at that level.
  3. 3
    Click Save
    The threshold takes effect within about ten seconds.
3. Check a waveform arrives
The next time the sensor reports a reading at or above the level, the app requests a waveform. To see one sooner, click Collect waveform now at the top of the sensor’s page: the request is sent the next time the sensor reports.
The capture appears under Recent captures as Requested, and changes to Stored when the waveform has arrived, usually within a minute. Click Download CSV to save it.
05Sensors
The Sensors screen lists every sensor the app watches. Click a row to open that sensor’s page.
ColumnShows
SensorThe name, the last bytes of the sensor’s address, 2 probes on a two-probe sensor, and what it is installed on and where.
Velocity, RMS acceleration, Acceleration maxThe latest value of each, taken as the largest of the X, Y and Z axes. A two-probe sensor shows one line per probe, marked P1 and P2. A value is teal when it is at or above its Waveform Collection level and red at or above its Critical level.
Collect at mostThe sensor’s minimum time between collections: Every crossing, or Once per a time, with when the next collection becomes possible.
Last waveformWhen the last capture was requested or stored, and why.
Last reportHow long ago the sensor last reported. A sensor that is not reporting cannot trigger anything.
StateBelow levels, Collecting, Critical, Paused, or No thresholds when nothing is switched on.
06Sensor Page
Everything about one sensor: what it collects, when, and its thresholds.
assetpulse lite atrium app guide 03 sensor settings
Sensors › a two-probe sensor
Summary and buttons
ButtonWhat it does
Collect waveform nowRequests a waveform whatever the thresholds say. The request is sent the next time the sensor reports. While a request is outstanding the button reads Waveform requested….
EditChanges the Name, Installed on, Location, Notes and Critical alert recipients for this sensor. Leave the recipients empty to use the default list in Settings.
Pause / ResumeStops every threshold on the sensor without deleting them. Readings are still shown.
Remove, then Confirm removeRemoves the sensor from the app with its thresholds, capture log and alert history. Waveforms already stored on the gateway are kept.
The summary shows the sensor’s address and type, what it is installed on, how long ago it last reported with its report interval and battery, the last collection, and who Critical alert emails go to.
Limits how often the Waveform Collection threshold collects. It is set per sensor, in hours and minutes, from 0 to 30 days. The default is 0.
SettingBehavior
0 (Every crossing)One waveform each time a level is crossed. The threshold is ready again once the reading drops back below the level.
Any other time (Once per)While a reading is at or above its level, a waveform is collected at most once in that time. A motor that starts several times a day and has 12 h set is collected from twice a day.
A Critical threshold ignores this limit. On a two-probe sensor the time is shared by both probes: a collection triggered by one probe also holds off the other.
Waveform request
What is collected when a threshold on this sensor fires.
FieldOptionsNotes
ProbesBoth probes (default), Only the probe that crossed its level, Always probe 1, Always probe 2Shown on two-probe sensors only. See Two-probe sensors.
CommandStandard (default), CustomStandard uses the waveform length and sample rate set on the sensor itself. Custom sets them for each request.
Sample rate100, 200, 400, 800, 1,600, 3,200, 6,400, 12,800 or 25,600 samples per secondCustom only.
Duration0.05 to 5 seconds, in steps of 0.05Custom only. The number of samples per axis is shown beside it.
The sensor also limits how many samples it will send, set by Set Max Raw Sample on the gateway’s sensor configuration page (1,024 to 8,100). A custom request for more than that returns the sensor’s limit.
Thresholds
Each of four metrics has its own panel, with a chart of its recent history and two independent thresholds. Use 24 h, 7 d or 30 d at the top right to change the charts’ range.
MetricUnitValue compared
Velocitymm/sLargest of the X, Y and Z axes
RMS accelerationgLargest of the X, Y and Z axes
Acceleration maxgLargest of the X, Y and Z axes
Temperature°C, or °F when the gateway is set to FahrenheitThe sensor’s temperature
assetpulse lite atrium app guide 05 single probe thresholds
Sensors › a sensor › Thresholds › Velocity
Waveform Collection
FieldDefaultNotes
SwitchOffCannot be switched on without a level.
Collect atemptyIn the metric’s unit. A waveform is collected when the reading is at or above it, as often as Minimum time between collections allows.
Critical
FieldDefaultNotes
Alarm atemptyIn the metric’s unit. With both thresholds on, it must be at or above the Waveform Collection level.
Collect a waveformTickedCollects even if a waveform was collected a moment ago.
Send an emailTickedTo the sensor’s own recipients, or the default list in Settings.
A Critical threshold acts once when the reading reaches the level, and not again until the reading has dropped back below the level and reached it again. Each time is listed on the Alerts screen.
Click Save in a panel to store that metric’s thresholds, or Discard to drop your changes. The dashed lines on the chart follow the levels as you type, so you can see where a level would sit before saving.
Two-probe sensors
A two-channel sensor is added once and appears as one sensor. Its name, minimum time between collections, alert recipients and waveform request are shared by both probes. Its thresholds are set separately for each probe.
Below the waveform request are two tabs, Probe 1 and Probe 2. Each tab shows that probe’s four metric panels, and carries a chip with where that probe stands.
assetpulse lite atrium app guide 04 probe thresholds
Sensors › a two-probe sensor › Probe 1 › Velocity
The Probes field of the waveform request decides which probe’s waveform a threshold collects:
ProbesWhat a threshold collects
Both probes (default)A threshold on either probe collects both with one command. They are logged as two captures, and each says which probe’s reading triggered it.
Only the probe that crossed its levelA threshold on a probe collects that probe only. If both cross on the same report, both are collected with one command.
Always probe 1 / Always probe 2A threshold on either probe collects that one probe.
07Captures
The log of every waveform requested for a watched sensor, newest first, with why each was collected. The same table appears on each sensor’s page as Recent captures.
assetpulse lite atrium app guide 07 captures
AssetPulse Lite › Captures
Filter the list by sensor, trigger and status with the three menus at the top.
ColumnShows
TimeWhen the waveform was stored, or requested if it has not arrived.
SensorThe sensor. Click it to open the sensor’s page.
TriggerWhy it was collected (below). On a two-probe sensor, Waveform of probe 1 or 2 says which probe the waveform is of.
ReadingThe reading that triggered it and the level it crossed. On a two-probe sensor it names the probe the reading came from.
StatusRequested, Stored or Failed, with the reason for a failure and a note when the command sent differs from the one configured.
WaveformThe number of samples per axis, the duration and the sample rate the sensor reported.
Download CSVSaves the waveform as a CSV file with a time column and one column per axis, in g.
Triggers
TriggerMeaning
Waveform CollectionA Waveform Collection threshold.
CriticalA Critical threshold.
ManualSomeone clicked Collect waveform now, or asked through the API.
OtherThe waveform arrived without this app asking: the sensor’s own schedule, another app, or the gateway’s sensor page.
The samples themselves are stored by the gateway and follow its data retention setting. A capture whose samples have since been removed reads No longer stored.
08Alerts
Every time a reading reached its Critical level, newest first. The Waveform Collection threshold never creates an alert.
assetpulse lite atrium app guide 08 alerts
AssetPulse Lite › Alerts
An alert opens when a reading reaches its Critical level and clears when the reading drops back below it. Use All, Open and Cleared to filter.
ColumnShows
OpenedWhen the reading reached the level.
Sensor, MetricWhich sensor and metric, and the probe on a two-probe sensor.
Reading, Critical levelThe reading that opened the alert, and the level at that time.
StateOpen, or when it cleared.
EmailWhether the alert email went out, Off when the threshold does not send one, or the reason it could not be sent.
WaveformWhether a waveform was requested for the alert.
09Settings
Settings that apply to every sensor. Each panel has its own Save button.
assetpulse lite atrium app guide 09 settings
AssetPulse Lite › Settings
FieldDefaultNotes
Default recipientsemptyComma-separated email addresses, used by every sensor that has no recipients of its own. Email goes through the mail server set under the gateway’s Settings › SMTP.
Send test emailSends a test message to the recipients, to confirm the mail server works.
One email is sent each time a reading reaches its Critical level. No further email is sent for that metric until it has dropped below the level and reached it again.
MQTT publishing
Publishes each collected waveform and each Critical alert through the gateway’s own MQTT client. The gateway must also grant the app access under its Settings › MQTT Client; until it does, the panel reads Not granted on the gateway.
FieldDefaultNotes
Publish each collected waveform and each Critical alert…OffSwitches publishing on.
Waveform topicatrium/::Gateway_Name::/assetpulse-lite/::Device_Id::/waveformMay contain the tokens listed under the fields.
Alert topicatrium/::Gateway_Name::/assetpulse-lite/::Device_Id::/alert
Include the samples in each waveform messageOnWith this off, a message carries the capture’s details and its id, and the samples are fetched over the REST API.
QoS10, 1 or 2.
RetainNo
Topic tokens: ::Gateway_Name::, ::Sensor_Name::, ::Device_Id:: (the address without colons), ::Probe::, ::Location::, ::Asset::, ::Metric:: and ::Trigger::.
Capture log
FieldDefaultNotes
Keep the capture log and cleared alerts for365 days1 to 3,650 days. This is the app’s log of why each waveform was collected. The samples are kept by the gateway under its own data retention setting.
10HTTP REST API
Everything the screens do is available to your own software over HTTP: adding sensors, setting thresholds, requesting waveforms and retrieving the samples.
All requests go to the gateway’s own address. Bodies and responses are JSON, keys are snake_case, and every time is in milliseconds since the Unix epoch. The examples use $GW for the gateway’s address and $TOKEN for an API token.
Create an API token
  1. 1
    Open API Tokens
    On the gateway dashboard, open Settings › Users & Security › API Tokens.
  2. 2
    Name it and choose its scopes
    Enter a Name and tick the scopes in the table below. Tick nothing else: a token should hold only what its software needs.
  3. 3
    Click Issue Token
    Copy the token and store it as you would a password. Send it with every request in an Authorization: Bearer header.
ScopeNeeded for
apps:dataEvery read in this section.
apps:actionsEvery change in this section, including requesting a waveform.
apps:readOptional. Lets the software list the gateway’s installed apps and what each offers.
Reading
Reads are GET requests under http://$GW/api/apps/assetpulse-lite.
RouteReturns
GET /sensorsEvery watched sensor: its id, address, name, probes, latest values per probe, the thresholds that are switched on, its state and its last capture.
GET /sensors/{sensor_id}The full setup of one sensor: all four metrics of each probe, including levels that are stored but switched off.
GET /available-sensorsThe vibration sensors on the gateway, their probes, and whether each is already watched.
GET /sensors/{sensor_id}/telemetryThe recent history of one metric of one probe. Query: probe (default 1), metric (default velocity), hours (default 24, up to 720), points (default 200). Each point is [bucket_start_ms, min, max, avg].
GET /capturesThe capture log, newest first. Query: sensor_id, probe, trigger, status, since, until, limit (default 50, up to 500), offset.
GET /captures/{capture_id}One capture’s details.
GET /captures/{capture_id}/waveformThe samples. See Retrieving waveforms.
GET /alertsCritical alerts, newest first. Query: sensor_id, open, since, limit.
GET /settingsThe default recipients, the MQTT settings and the log retention.
sensor_id is the app’s own number for a watched sensor, from GET /sensors. It is not the sensor’s radio address.
bash List the watched sensors
curl -s -H "Authorization: Bearer $TOKEN" \
  "http://$GW/api/apps/assetpulse-lite/sensors"
json Response (one two-probe sensor, shortened)
  "sensors": [
      "id": 2,
      "device_id": "00:13:a2:00:42:38:28:26",
      "probes": ["1", "2"],
      "name": "Fault Simulator Motor",
      "enabled": true,
      "min_interval_minutes": 0,
      "waveform_probes": "both",
      "waveform_command": "standard",
      "sensor_type": 111,
      "thresholds": [
        { "probe": "1", "metric": "velocity", "collect_enabled": true, "collect_level": 1,
          "critical_enabled": false, "critical_level": null, "state": "collecting" },
        { "probe": "2", "metric": "velocity", "collect_enabled": true, "collect_level": 1,
          "critical_enabled": true, "critical_level": 8, "state": "collecting" }
      "values": {
        "1": { "velocity": 2.28, "rms_acc": 0.127, "max_acc": 0.071 },
        "2": { "velocity": 4.15, "rms_acc": 0.169, "max_acc": 0.095 }
  "count": 1
A sensor’s probes is ["1"], or ["1", "2"] for a two-probe sensor. Every threshold carries the probe it is on, and values are keyed by probe. A threshold’s state is ok, collecting or critical.
Configuring
Changes are made by running one of the app’s six actions. Each is a POST to http://$GW/api/bus/actions/assetpulse-lite/{action} with a JSON body of {"params": {...}}.
ActionParameters
sensor.adddevice_id (required: the sensor’s full address, or its last bytes such as 42:38:30:e2), name, location, asset, notes, min_interval_minutes, and the waveform parameters below.
sensor.updatesensor_id (required), name, location, asset, notes, enabled, min_interval_minutes, notify_recipients, and the waveform parameters below.
sensor.removesensor_id (required). Removes the sensor’s thresholds, capture log and alert history. Stored waveforms are kept.
threshold.setsensor_id and metric (both required), probe (required on a two-probe sensor), collect_enabled, collect_level, critical_enabled, critical_level, critical_waveform, critical_notify.
waveform.capturesensor_id (required). Requests a waveform whatever the thresholds say. The waveform parameters, if given, apply to this one request only.
settings.updatenotify_recipients, mqtt_enabled, mqtt_waveform_topic, mqtt_alert_topic, mqtt_include_samples, mqtt_qos, mqtt_retain, capture_log_retention_days.
metric is velocity, rms_acc, max_acc or temperature. Levels are in the metric’s unit (mm/s, g, g and °C). min_interval_minutes is 0 to 43200. Yes/no parameters take true or false.
ParameterValues
waveform_probesTwo-probe sensors: both (default), triggering (only the probe that crossed its level), 1 or 2. A single-probe sensor accepts only 1.
waveform_commandstandard (default) or custom.
waveform_sample_rateRequired for custom: 100, 200, 400, 800, 1600, 3200, 6400, 12800 or 25600 samples per second.
waveform_duration_msRequired for custom: 50 to 5000, in steps of 50.
bash Watch a sensor, then collect when velocity reaches 2 mm/s
curl -s -X POST -H "Authorization: Bearer $TOKEN" -H "Content-Type: application/json" \
  "http://$GW/api/bus/actions/assetpulse-lite/sensor.add" \
  -d '{"params": {"device_id": "00:13:a2:00:42:38:30:e2", "name": "Pump 3 drive end",
       "asset": "Pump 3", "location": "Pump room", "min_interval_minutes": 720}}'
  "http://$GW/api/bus/actions/assetpulse-lite/threshold.set" \
  -d '{"params": {"sensor_id": 1, "metric": "velocity",
       "collect_enabled": true, "collect_level": 2,
       "critical_enabled": true, "critical_level": 8}}'
An action answers with ok and a detail string that is itself JSON: what is now stored. Parse detail to read it.
json Response to threshold.set
  "ok": true,
  "detail": "{\"probe\":\"1\",\"metric\":\"velocity\",\"label\":\"Velocity\",\"unit\":\"mm/s\",\"readings\":[\"x_velocity_mm_sec\",\"y_velocity_mm_sec\",\"z_velocity_mm_sec\"],\"collect_enabled\":true,\"collect_level\":2,\"critical_enabled\":true,\"critical_level\":8,\"critical_waveform\":true,\"critical_notify\":true,\"state\":\"ok\",\"value\":0.31,\"sensor_id\":1}",
  "duration_ms": 6
sensor.add and sensor.update return the sensor’s settings and the configuration of each threshold that is set. For live values and each threshold’s state, read GET /sensors/{sensor_id}.
Two-probe sensors
A two-probe sensor is one sensor_id. What it collects is set once on the sensor, and thresholds are set per probe by adding probe to threshold.set. On a two-probe sensor probe is required; a call without it is refused.
bash Collect both probes with a custom command when either probe passes 2 mm/s
  "http://$GW/api/bus/actions/assetpulse-lite/sensor.update" \
  -d '{"params": {"sensor_id": 2, "waveform_probes": "both", "waveform_command": "custom",
       "waveform_sample_rate": 3200, "waveform_duration_ms": 1000}}'
for PROBE in 1 2; do
    -d "{\"params\": {\"sensor_id\": 2, \"probe\": \"$PROBE\", \"metric\": \"velocity\",
         \"collect_enabled\": true, \"collect_level\": 2}}"
done
One command then collects both probes. The capture log holds two captures, one per probe. Each has the probe the waveform is of and the trigger_probe whose reading triggered the request.
bash Request both probes now, without changing the sensor’s settings
  "http://$GW/api/bus/actions/assetpulse-lite/waveform.capture" \
  -d '{"params": {"sensor_id": 2, "waveform_probes": "both"}}'
waveform.capture returns capture_ids, one per probe. The request is sent the next time the sensor reports. Asking again while a request is outstanding returns the same request instead of queuing another.
List the captures that have arrived, then fetch each one’s samples by its id.
bash Stored captures of one sensor
  "http://$GW/api/apps/assetpulse-lite/captures?sensor_id=2&status=stored&limit=10"
json Response (one capture shown)
  "captures": [
      "id": 53, "sensor_id": 2, "name": "Fault Simulator Motor",
      "device_id": "00:13:a2:00:42:38:28:26", "probe": "2", "dual_probe": true,
      "trigger": "collection", "metric": "velocity", "trigger_probe": "2",
      "value": 6.84, "level": 1,
      "status": "stored", "requested_at": 1795556488000, "stored_at": 1795556523000,
      "raw_id": 35, "waveform_available": true,
      "samples": 3200, "odr_hz": 3258, "duration_s": 0.982,
      "command": "standard"
  "count": 1, "total": 1, "limit": 10, "offset": 0
FieldMeaning
probeThe probe the waveform is of.
triggercollection, critical, manual or other.
metric, value, level, trigger_probeThe reading that triggered the request, the level it crossed, and the probe it came from.
statusrequested, stored or failed. A failed capture has an error.
samples, odr_hz, duration_sWhat the sensor returned: samples per axis, the sample rate, and the length in seconds.
command, noteThe command that was sent. A custom command also has requested_sample_rate and requested_duration_ms. note appears when the command sent differs from the one configured.
waveform_availablefalse once the gateway’s data retention has removed the samples.
bash The samples of capture 53
  "http://$GW/api/apps/assetpulse-lite/captures/53/waveform"
json Response (sample arrays shortened)
  "capture_id": 53, "raw_id": 35, "sensor_id": 2,
  "device_id": "00:13:a2:00:42:38:28:26", "trigger": "collection",
  "captured_at": 1795556523026, "odr_hz": 3258, "samples": 3200,
  "units": { "acceleration": "g", "time": "s" },
  "probe": "2",
  "x": [0.012, -0.004],
  "y": [0.001, 0.003],
  "z": [-0.008, 0.002]
x, y and z are acceleration in g. The time of sample i is i / odr_hz seconds. A waveform is typically thousands of samples per axis; pass offset and limit to fetch part of it.
Errors and limits
ResponseMeaning
401No token was sent, or it is not valid.
403The token’s scopes do not cover the route. Only the reads listed above are open to a token; the app’s own screens use other routes that are not.
400 from a readA parameter is not valid. The body is {"error": "..."} and says which.
400 from an actionThe action was refused. The body is {"ok": false, "code": "action_failed", "error": "..."}, and error says why, for example that a two-probe sensor needs probe.
404No sensor or capture has that id.
409 from a waveformThe capture has not arrived yet, or it failed.
410 from a waveformThe gateway’s data retention has removed the samples.
LimitValue
An action’s detail2,000 characters. Every action’s result fits. On a sensor with a very long note, sensor.update leaves the thresholds out and says so in a note.
Clearing a text fieldAn action cannot clear a name, location, asset or note, because an empty parameter means “keep”. Send a new value, or clear it on the sensor’s page.
CSV downloadAvailable from the screens. Software with a token fetches the samples as JSON.
11Integration
What the app offers the rest of the gateway: an event and actions for Automations, tools for an AI assistant, and MQTT messages.
Automations
The app publishes one event on the gateway’s Automations page, Vibration metric crossed a level (assetpulse-lite/threshold.state), once per sensor, probe and metric. It can be wired to any action on the gateway, such as a relay, and releases by itself when the reading recovers.
FieldValues
stateok, collecting (at or above the Waveform Collection level) or critical.
collection_breachedtrue whenever the reading is at or above the Waveform Collection level, including while critical. Use it for “the machine is running”.
critical_breachedtrue while the reading is at or above the Critical level. Use it for alarms.
Alsosensor_id, device_id, probe, name, location, asset, metric, unit, value, collect_level, critical_level.
All six actions in the HTTP REST API section can also be the “Then” of an automation.
AI assistants
An assistant connected to the gateway over MCP sees every read as a tool named assetpulse-lite__ followed by the read (sensors, sensor_setup, available_sensors, telemetry, captures, capture, capture_waveform, alerts, settings), and every action the same way with the dot replaced by a dash, such as assetpulse-lite__threshold-set. Its token needs apps:data to read and apps:actions to change anything.
MQTT messages
With MQTT publishing on, the app publishes a JSON message to the waveform topic for each stored waveform, and to the alert topic when a Critical alert opens and when it clears.
MessageFields
Waveformgateway, sensor_id, device_id, probe, name, location, asset, capture_id, raw_id, trigger, metric, unit, value, level, odr_hz, samples, temperature_c, captured_at, and with samples included, units and raw_data (x, y, z).
Alertgateway, sensor_id, device_id, probe, name, location, asset, alert_id, state (critical or cleared), metric, unit, value, critical_level, ts.
12Troubleshooting
Problems seen on real gateways, what caused them, and what fixed them.
SymptomCauseFix
Captures fail with No waveform received from the sensor, and the sensor’s settings keep changing.Two gateways are managing the same sensor.Leave the sensor managed by one gateway. On the others, switch off management of that sensor.
Every request to a sensor fails on a newly installed gateway.The gateway cannot reach the sensor: a missing or loose radio antenna, or Sensor Link Assurance switched off.Check the antenna, and switch on Sensor Link Assurance on the gateway.
A capture’s note reads Standard waveform requested: the sensor applies a custom command only in Smart operation mode.The sensor’s Mode setting is not Smart.Set Mode to Smart on the gateway’s sensor configuration page, or use the Standard command.
A custom capture returns fewer samples than requested.The sensor’s Set Max Raw Sample setting is lower than the request.Raise Set Max Raw Sample on the gateway’s sensor configuration page (up to 8,100).
On a two-probe sensor, probe 1’s waveform arrives and probe 2’s capture fails.The sensor was reset during the 30 seconds between the two waveforms, or its second probe is not answering. A sensor whose reports show a probe error cannot return that probe’s waveform.Leave the sensor alone until both captures settle. If it repeats, check the probe and its cable.
A large waveform fails with waveform incomplete: N of M packets.Some of the waveform’s radio packets were lost. Larger waveforms take longer to send and are more exposed.Improve the sensor’s radio link, or request fewer samples.
Nothing is collected although the machine is running.The sensor has not reported while the reading was above the level, the threshold is switched off, the sensor is paused, or the minimum time between collections has not passed.Check Last report, the threshold’s switch, the sensor’s state chip and Collect at most on the Sensors screen.
Captures stop for about an hour after several failures.After three requests in a row with no waveform, the app pauses automatic requests to that sensor for 60 minutes to save its battery.Fix the cause of the failures. Collect waveform now still works during the pause.
The Alerts screen shows Not sent: SMTP is not configured on this gateway.The gateway has no mail server set.Set one under the gateway’s Settings › SMTP, then use Send test email.
An API call returns 403.The token lacks apps:data or apps:actions, or the route is not one of the reads open to a token.Edit the token’s scopes, and use the routes in the HTTP REST API section.
13Reference
Timings, limits and the levels’ meaning in one place.
Levels
Both levels are values you choose for your own machine. They are not taken from a vibration severity standard such as ISO 10816 or ISO 20816, and the app does not grade readings against one.
Timings
ItemValue
A saved setting takes effectWithin about 10 seconds
A threshold request is sentRight after the report that crossed the level, inside the sensor’s listening window of about 2 seconds
A manual request is sentThe next time the sensor reports, and held for up to 24 hours
A second request to the same sensorNot sent while one from the last 90 seconds is outstanding
A request with no waveform is marked FailedAfter 3 minutes
Probe 2 of a both-probes requestAbout 30 seconds after probe 1
Pause after repeated failures60 minutes after 3 failed requests in a row
Limits
ItemLimit
Minimum time between collections0 to 43,200 minutes (30 days)
Custom sample rate100 to 25,600 samples per second
Custom duration0.05 to 5 seconds, in steps of 0.05
Samples per waveformThe sensor’s Set Max Raw Sample setting, up to 8,100
Capture log retention1 to 3,650 days (default 365)
Captures per API request500
Name / Location, Installed on / Notes120 / 200 / 2,000 characters
CSV file
A downloaded waveform is named waveform_, the sensor’s address, the probe on a two-probe sensor, and the capture time. Its columns are time_s, x_g, y_g and z_g.