- Overview
- Description
- Getting Started
- Modes of Operation
- Appendix A
- Configuration Commands
- 1. Set Broadcast Transmission
- 2. Set ID and Delay
- 3. Set Destination Address
- 4. Set Power
- 5. Set PANID
- 6. Set Retries
- 7. Read Delay
- 8. Read Power
- 9. Read Retries
- 10. Read Destination Address
- 11. Read PANID
- 12. Enable Encryption
- 13. Disable Encryption
- 14. Set Encryption Key
- 15. Set Filtering
- 16. Set Data Rate
- 17. Set Time Series Flag
- 18. Set Reading Type
- 19. Read Filtering
- 20. Read Data Rate
- 21. Read Time Series Flag
- 22. Read Reading Type
- Appendix B
- Industrial Grade 3-axis Vibration Sensor with RMS, MAX and MIN acceleration in g
- Time Domain data transmission for Time/Frequency domain Analysis
- Vibration Range ±16g
- Noise Removal using Multiple Low pass filters and Over sampling
- Configurable Bandwidth: 58 Hz to 12.8 kHz
- Configurable Sampling Rate 400 Hz to 25.6 kHz
- Built-in Offset and Gravity vector removal filters
- Built-in 6-point calibration for 0-g Offset and Sensitivity adjustment
- Temperature output with an accuracy of ±1°C
- Small Form Factor (3×3 inch)
- Default Factory setting Restore option
- Real time battery status
- Wireless Industrial Machine Health Analysis
- Machine Fault detection
- Frequency Analysis of Vibration data
- Building and Structural Monitoring
Introducing NCD’s Enterprise Wireless Vibration Sensor, boasting up to a 28 Mile range using a 900MHz wireless mesh networking architecture or 1 Mile using 2.4GHz wireless mesh networking architecture. Incorporating an DC coupled acceleration sensor and a temperature sensor, the frequency response is from 0 Hz to user defined upper cutoff frequency. It samples and processes vibration data at user defined sampling rate and bandwidth, and then transmits this information to user defined receiver(s). The whole process is repeated at user defined intervals.
The Enterprise Wireless Vibration Sensor removes the offsets and accounts for sensitivity variation of the internal sensing element. It has an internal memory which stores the 0-g offsets and sensitivities for each axis. Default factory saved values usually provide the required accuracy, however, the sensor incorporates a 6-point calibration routine that can be initiated in the configuration mode by the user to carry out 6-point calibration as described in configuration section. This process overrides the factory default values however, the settings can be restored as discussed in troubleshooting section.
At power up or reset, the sensor calibrates itself so that it can remove 0-g offsets as well as gravity vector from future readings. This process may take up 10 seconds. This process does not require the sensor to be placed on a stationary platform. Hence, it can be mounted on running machines at any angle and a simple reset of the sensor allows it to calibrate itself.
The sensor accumulates 340 samples for each of the 3-axis and processes them to calculate RMS, Maximum and Minimum Vibration in this interval. The sample accumulation period thus varies with the configurable sampling rate. A low sampling rate increases the sample accumulation period, hence care should be exercised since it increases the duration for which the sensor is powered up and thus drains the battery. For example, sample accumulation period for a sampling rate of 400 Hz turns out to be 850msec, whereas, for a sampling rate of 25.6 kHz, it is reduced to13.3msec. The resultant information is then combined with temperature read from a temperature sensor and sent to one or multiple receivers. After data transmission, the sensor goes to sleep for a user defined interval, thus minimizing power consumption.
Powered by just 2 AA batteries and an operational lifetime of 500,000 wireless transmissions, a 5 years battery life can be expected depending on environmental conditions and the data transmission interval. Optionally, this sensor may be externally powered.
With an open communication protocol this sensor can be integrated with just about any control system or gateway. Data can be transmitted to a PC, a Raspberry Pi, to Microsoft Azure® IoT, or Arduino. Sensor parameters and wireless transmission settings can be changed on the go using the open communication protocol providing maximum configuration depending on the intended application.
The range, price, accuracy, battery life and security features of Enterprise Wireless Vibration Sensor makes it an affordable choice which exceeds the requirements for most of the Industrial as well as consumer market applications.
The Vibration Sensors and Zigmo/Router come pre-programmed and work out of the box. In this section we will setup a sensor and Zigmo link and start receiving data on our PC. Though this guide shows how to visualize data on LabVIEW utility, you can also use a simple serial terminal to see raw data by following these steps.
- Industrial IoT Wireless Vibration & Temperature Sensor V2 MEMS (with power source Battery Or External DC)
- LabVIEW Utility for displaying sensor data (available on our website for free)
Note: The Wireless Sensor comes with external power enabled, for battery conservation during shipping. To enable battery power, open the enclosure and set the PS (power select) jumper which is parallel to the marking line on the board.
- Press the Run button to visualize incoming data.
Modes of Operation
- Run Mode
- Configuration Mode
Standard Data Transmission
Standard data transmission mode is the default mode of operation of this sensor. In this mode the sensor sends periodic packets to destination receiver. During the time it is not sending packets, it sleeps and conserves power. Sensor end X-bee operates in API mode and sends packets to the saved destination address on the network specified by the saved PAN ID. Figure 12 illustrates an API packet transmission and reception.
Typical response from the device in Standard data transmission mode is shown in Figure 13 and Figure 14. The utility shown in Figure 14 can be downloaded from the website.
Frame Field | Offset (Payload Section) | Fixed Value (if any) | Description | ||||||||||||||||||
Header | 0 | 0x7F |
Firmware | 2 | – |
Battery Voltage | MSB 3 | – | Sampled battery voltage of the device. Battery Voltage=((Battery Voltage MSB x 256+Battery Voltage LSB) x 0.00322 V | ||||||||||||||||||
LSB 4 | – | ||||||||||||||||||||
Packet Counter | 5 | – |
Sensor Type | MSB 6 | 0x00 |
MSB 7 | 0x28 | ||||||||||||||||||||
Status (b :bit) | 8 | 0x00 | Sensor Status as defined below, shift the status bit right 2 places for the integer value (b >> 2), the value in parentheses is there for completeness and is consistent with the data sheet.
b1 | Data Type in Payload defined by following: 1: Acceleration (g) 2: Velocity (inch/sec) 3: Time domain data (acceleration in g) | ||||||||||||||||||||
b0(LS) | |||||||||||||||||||||
Vibration RMS | 9/X[0] | – | Vibration RMS X-axis Data (24 bit Signed Output) Vibration (mg) = ((X[0]<<16)+(X[1]<<8)+X[2])/100 (Saving directly to 32bit integer will give wrong result | ||||||||||||||||||
10/X[1] | – | ||||||||||||||||||||
11/X[2] | – | ||||||||||||||||||||
12/Y[0] | – | Vibration RMS Y-axis Data (24 bit Signed Output) Vibration (mg) = ((Y[0]<<16)+(Y[1]<<8)+Y[2])/100 (Saving directly to 32bit integer will give wrong result | |||||||||||||||||||
13/Y[1] | – | ||||||||||||||||||||
14/Y[2] | – | ||||||||||||||||||||
15/Z[0] | – | Vibration RMS Z-axis Data (24 bit Signed Output) Vibration (mg) = ((Z[0]<<16)+(z[1]<<8)+Z[2])/100 (Saving directly to 32bit integer will give wrong result | |||||||||||||||||||
16/Z[1] | – | ||||||||||||||||||||
17/Z[2] | – | ||||||||||||||||||||
Vibration Max | 18/X[0] | – | Vibration Max X-axis Data (24 bit Signed Output) Vibration (mg) = ((X[0]<<16)+(X[1]<<8)+X[2])/100 (Saving directly to 32bit integer will give wrong result | ||||||||||||||||||
19/X[1] | – | ||||||||||||||||||||
20/X[2] | – | ||||||||||||||||||||
21/Y[0] | – | Vibration Max Y-axis Data (24 bit Signed Output) Vibration (mg) = ((Y[0]<<16)+(Y[1]<<8)+Y[2])/100 (Saving directly to 32bit integer will give wrong result | |||||||||||||||||||
22/Y[1] | – | ||||||||||||||||||||
23/Y[2] | – | ||||||||||||||||||||
24/Z[0] | – | Vibration Max Z-axis Data (24 bit Signed Output) Vibration (mg) = ((Z[0]<<16)+(Z[1]<<8)+Z[2])/100 (Saving directly to 32bit integer will give wrong result | |||||||||||||||||||
25/Z[1] | – | ||||||||||||||||||||
26/Z[2] | – | ||||||||||||||||||||
Vibration Min | 27/X[0] | – | Vibration Min X-axis Data (24 bit Signed Output) Vibration (mg) = ((X[0]<<16)+(X[1]<<8)+X[2])/100 (Saving directly to 32bit integer will give wrong result | ||||||||||||||||||
28/X[1] | – | ||||||||||||||||||||
29/X[2] | – | ||||||||||||||||||||
30/Y[0] | – | Vibration Min Y-axis Data (24 bit Signed Output) Vibration (mg) = ((Y[0]<<16)+(Y[1]<<8)+Y[2])/100 (Saving directly to 32bit integer will give wrong result | |||||||||||||||||||
31/Y[1] | – | ||||||||||||||||||||
32/Y[2] | – | ||||||||||||||||||||
33/Z[0] | – | Vibration Min Z-axis Data (24 bit Signed Output) Vibration (mg) = ((Z[0]<<16)+(Z[1]<<8)+Z[2])/100 (Saving directly to 32bit integer will give wrong result | |||||||||||||||||||
34/Z[1] | – | ||||||||||||||||||||
35/Z[2] | – | ||||||||||||||||||||
Temperature Data | 36/T[0] | – | Temperature data (16 bit Signed Output) Temperature (°C) = (T[0]<<8)+T[1] | ||||||||||||||||||
37/T[1] | – |
Time domain data transmission
The Enterprise Wireless Vibration Sensor enables time domain data to be downloaded from the sensor. This enables a vibration analyst to dig deep into the root cause by processing time domain data and analyzing it in both time and frequency domains. The sampling rate and bandwidth can be configured according to the frequency bin resolution and the range of frequencies to be analyzed in the frequency analysis.
The time domain data can only be queried using the command(in hex) shown below:
7E00 1310 0000 0000 0000 00FF FFFF FE00 00F8 9900 0028 3B
As mentioned earlier, the sensor goes back to sleep after successfully sending standard data transmission packet. However, if TIME SERIES flag is enabled from the configuration mode, the sensor waits an additional 600msec to wait for the time series command. If the above command is received during this interval, it sends out time domain data and goes back to sleep. Note that, the sensor power consumption is increased if time series flag is enabled.
Time domain data comprises of 340 samples for each axis. Each sample is a 16-bit integer. Hence, total time domain data contains 2040 bytes. Due to the limitation of packet size for X-bee communication, this data is segmented and sent in 12 packets as shown in following figure. Data has to be recombined at the receiver end into 340 samples for each axis before further processing. It should be noted that since the amount of data is big therefore it is only sent on demand and at the destination address saved inside the sensor. For this purpose, the source address from the packet containing command is matched with the saved destination address. This step essentially verifies that the command is coming from designated node. If, however, the destination address stored inside the sensor is a broadcast address, then this matching is disabled and the time domain data is sent to all listening nodes.
The 2 byte data samples are integer counts of the acceleration and must be converted into acceleration unit(g) by dividing the count value with 2048 since 1 count equals 0.488mg. The following Figure shows a User Interface that display time domain data on the left side and its Fast Fourier Transform (FFT) on the right side.
RMS Noise for various bandwidth settings
Following figure illustrates the RMS noise present in sensor acquired data against various bandwidth settings.
A standard configuration packet and its fields are explained in Figure 18. Its possible responses are also shown. The commands supported by this sensor are shown in Table 5, these can be used in the Parameters field of Payload section. The sensor responds to these commands with an acknowledgement if the process completed successfully or with an error if it failed to setup a parameter. The respective Data and Reserve section length and values are shown in Table 6 for the case of acknowledgement. In the case of error, the reserved section will be fixed and not used, while the Error number byte will determine the type of error returned. These errors are mentioned in Table 7.
Figure 18 depicts standard communication between Zigmo/Router and sensor. Sensor commands have variable length frames whereas responses received from sensor are fixed length. The 2 scenarios are also shown, where a command can result in an acknowledgement reception or an error reception at the Zigmo end.
The Configuration mode also supports 6-point calibration to improve the accuracy of the results by determining 0-g offsets and sensitivities for each axis. The command to initiate 6-point calibration is provided in the table below. The response received from the sensor is in ASCII text and self explanatory. To advance the calibration process, GO command is required to be sent to the sensor which is also described in the table below.
Examples for setting parameters in configuration mode are shown in Appendix A. Note that incorrectly setting some of the critical settings such as setting PANID and Destination address can disable further communication between a Zigmo and the sensor. Hence, these settings are saved in the sensor but they take affect after a reset so that the communication is not lost.
The sensor has 4 time domain data modes
1. No time domain
sample complete API packet
7E 00 14 10 00 00 00 00 00 00 00 FF FF FF FE 00 00 F4 08 00 00 28 01 CF
Command — F4 08 00 00 28 01
2. Only time domain data
7E 00 14 10 00 00 00 00 00 00 00 FF FF FF FE 00 00 F4 08 00 00 28 02 CE
Command — F4 08 00 00 28 02
3. Time domain data with each packet
7E 00 14 10 00 00 00 00 00 00 00 FF FF FF FE 00 00 F4 08 00 00 28 03 CD
Command — F4 08 00 00 28 03
4. Time domain on demand (This command needs to be sent within 1 sec of data reception)
7E 00 14 10 00 00 00 00 00 00 00 FF FF FF FE 00 00 F4 08 00 00 28 04 CC
Command — F4 08 00 00 28 04
Once the gateway receives the data, it needs to send a command to request on demand time domain data
7E 00 13 10 01 00 00 00 00 00 00 FF FF FF FE 00 00 F8 99 00 00 28 3A
Command — F8 99 00 00 28
F8 — Command Header
99 — Sub command
00 — node id
00,28 — Sensors type
Once the sensor gets this command it will send RMS, Max,Min, vibration data plus the time domain data
The UI, shown in Figure 19, can be used to configure the wireless sensor. At startup, it automatically changes the Zigmo PANID so that it can communicate with a sensor in configuration mode (indicated by the Led on the top right). Upon exit, the PANID of the Zigmo is restored to old value.
Individual settings can be programmed using the single command column.
The AUTO PROGRAMMING check option allows the user to setup multiple sensors with the same settings. In such a scenario, the user will be required to enable each check box for enabling a setting, enter the value for each setting if required and then check the auto programming check box. Afterwards, when a sensor is powered up and enters configuration mode, the PGM MODE DETECTED Led will flash and automatically program the checked settings. User can also program multiple settings by clicking the APPLY SELECTED button. Moreover, settings can be read using the individual buttons or all settings can be read using the READ ALL button.
Appendix A
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F7 0100 0028 D4
Command For COPY: 7E00 1710 0000 0000 0000 00FF FFFF FE00 00F7 0200 0028 0000 0005 CE
Command For COPY: 7E00 1710 0000 0000 0000 00FF FFFF FE00 00F7 0300 0028 1234 5678 BE
Command For COPY: 7E00 1410 0000 0000 0000 00FF FFFF FE00 00F7 0400 0028 03CE
Command For COPY: 7E00 1510 0000 0000 0000 00FF FFFF FE00 00F7 0500 0028 7CDE 76
Command For COPY: 7E00 1410 0000 0000 0000 00FF FFFF FE00 00F7 0600 0028 04CB
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F7 1500 0028 C0
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F7 1600 0028 BF
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F7 1700 0028 BE
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F7 1800 0028 BD
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F7 1900 0028 BC
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F2 0100 0028 D9
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F2 0200 0028 D8
Command For COPY: 7E00 2410 0000 0000 0000 00FF FFFF FE00 00F2 0300 0028 0011 2233 1122 3311 2233 1122 3311 2233 4495
15. Set Filtering
Command For COPY: 7E00 1410 0000 0000 0000 00FF FFFF FE00 00F4 0200 0028 01D5
16. Set Data Rate
Command For COPY: 7E00 1410 0000 0000 0000 00FF FFFF FE00 00F4 0300 0028 0FC6
Command For COPY: 7E00 1410 0000 0000 0000 00FF FFFF FE00 00F4 0800 0028 01CF
Command For COPY: 7E00 1410 0000 0000 0000 00FF FFFF FE00 00F4 0400 0028 01D3
19. Read Filtering
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F4 0500 0028 D3
20. Read Data Rate
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F4 0600 0028 D2
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F4 0900 0028 CF
Command For COPY: 7E00 1310 0000 0000 0000 00FF FFFF FE00 00F4 0700 0028 D1
Appendix B
- Add all the bytes and keep the lower 8 bits of result (Excluding the frame delimiter and length)
- Append this byte at the end of the original packet for sending
- Add all the bytes including the received checksum (Exclude the frame delimiter and length)
- Keep only the last 8 bits